Warning about ICP procedures that can cause false readings: freezing

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
72,100
Reaction score
69,741
Location
Massachusetts, United States
Rating - 0%
0   0   0
I do not know what tube filling procedures are advised by all ICP companies, but a recent test by Christoph at Oceamo points out a potentially serious concern.

Freezing of seawater causes the formation of fresh water ice and hypersaline seawater. As the ice spreads to more and more of the total water, the saline liquid remaining can be very much concentrated relative to the original sample.

If all of the collected sample remains within the tube, there will be no change the in the element concentrations when the sample is thawed (at least not by the mechanism described below).

But if the tube is filled to near or at the top, having little air volume to compress, the pressure can rise quite substantially during freezing, and this rise can push out some of the hypersaline liquid. Since that liquid no longer remains in the tube when the sample is thawed, those ions are lost from the measurement, and the net concentration of the remaining fluid after thawing is reduced. Christoph showed that a tube filled to the top and frozen to -4 deg F showed a drop in calcium from about 420 ppm to about 300 ppm, magnesium from about 1250 ppm to 900 ppm, etc. he also showed very high ion concentrations in the exuded fluid

These effects are far too large to ignore.

Thus, if there is any chance your samples may freeze, be sure to NOT fill them all the way up. If your ICP company advises filling to the top, I'd suggest contacting them to see if it is OK to only half fill the tubes.

Here is Christoph's data:

 

dschuffert

Well-Known Member
View Badges
Joined
Aug 11, 2022
Messages
669
Reaction score
1,888
Location
USA
Rating - 0%
0   0   0
Thanks for the information.
 

Christoph

Active Member
View Badges
Joined
May 23, 2017
Messages
273
Reaction score
592
Location
Vienna, Austria
Rating - 0%
0   0   0
Hi Randy,

thanks for posting this in prominent position!

@all: Please do not fill the vials only halfway full - the labs will run into issues because of limited water volume. Having a air cusion of ~10% is sufficient. So fill a 15 ml vial only to the 13-14 ml and a 30 ml vial only to the ~27 ml mark. This is sufficient to prevent any leaking, which will occur when the vials are filled completely.

@RHF: Is there a way dropping you a PM?

Thanks and best regards,
Christoph
 

rtparty

5000 Club Member
View Badges
Joined
Jan 19, 2010
Messages
5,388
Reaction score
9,137
Location
Utah
Rating - 100%
4   0   0
A hot pack lasts like 24-48 hours tops. Most of these tests take 7-10 days to get where they’re going. The hot pack won’t make the trip
 

PSXerholic

Valuable Member
View Badges
Joined
Nov 3, 2015
Messages
2,010
Reaction score
3,204
Location
Houston
Rating - 0%
0   0   0
Very good point, I do communicate this in the RM Facebook Group during the Winter season for the Reef Moonshiner's Testkits (Oceamo ICP-OES/MS).

Time to implement this detail into the Sampling Instructions.
Great work from Christoph and the Oceamo Team, thanks for sharing Randy!

1706130818193.png
 

Suspect

Community Member
View Badges
Joined
May 17, 2016
Messages
80
Reaction score
67
Location
Wisconsin
Rating - 100%
1   0   0
I sent in an ATI ICP-OES test a couple weeks ago when it was cold. Salinity, Kh, Ca, and Mg were all suspiciously low compared to my Trident and hand tests. ATI commented that the RODI sample leaked so much they were unable to perform the RODI test. The samples must have frozen. I've always filled the tubes completely because I remember reading somewhere that the sample can interact with air.
 

CaptainOver

Community Member
View Badges
Joined
Dec 20, 2016
Messages
86
Reaction score
100
Location
Seattle
Rating - 0%
0   0   0
Maybe a dumb question, but if the tube is upside down when it freezes (I doubt it can be kept vertical while shipping), won't some liquid be forced out when the air at the top expands/contracts? Or is the cap seal tight enough that nothing is forced out and the 10% of air just absorbs the pressure change?
 

dschuffert

Well-Known Member
View Badges
Joined
Aug 11, 2022
Messages
669
Reaction score
1,888
Location
USA
Rating - 0%
0   0   0
Thanks for the info.
 

Christoph

Active Member
View Badges
Joined
May 23, 2017
Messages
273
Reaction score
592
Location
Vienna, Austria
Rating - 0%
0   0   0
Maybe a dumb question, but if the tube is upside down when it freezes (I doubt it can be kept vertical while shipping), won't some liquid be forced out when the air at the top expands/contracts? Or is the cap seal tight enough that nothing is forced out and the 10% of air just absorbs the pressure change?
The tubes can withstand quite some pressure. The air in the tube just allows the liquid to expand without leaking, regardless of orientation of the tubes.

All the best, Christoph
 