SAMPLE STORAGE AND ITS IMPACT ON PHOSPHATE MEASUREMENT
In a discussion on R2R @BigJohnny posed a question about why his Hanna ULR Phosphorous test results were significantly higher (2-3 times) than his ICP test results…. Here is a link to the discussion
Accuracy of Phosphate ICP analysis
Curious because my Hanna ulr phosphorus shows 2-3x more on average than icp. I stopped using it though because my experience with hanna is that they are consistent but accuracy/reagents can vary so it's better for relative measurements, and that assumes perform the test perfectly and clean the...
www.reef2reef.com
One of the possible reasons that were proposed was the potential impact of sample transit time and could that have an impact on the results of the ICP test. So I set up a series of experiments to test the hypothecs. At first it was a simple one time experiment in which I took a large sample and measured the dKH, NO3, PO4 and Iodine Levels on the initial sampling. Each was measured 3 times and an average value calculated. The dKH was measured with the Hanna Checker. The Nitrate was measured using the test I developed using the Hanna HI-764 (link to details--- https://www.reef2reef.com/ams/test-meter-for-testing-nitrates.599/ )
The PO4 was tested using the Hanna HI-736 Phosphorous ULR Checker and the Iodine was tested using the method I developed ( link to details https://www.reef2reef.com/ams/using-hanna-checker-hi-707-to-test-for-iodine.743/ )
I took the rest of the remaining sample and split into two sample sets. The first one I stored inside in a controlled environment. The second one I placed in the glove compartment of my car…Uncontrolled environment. I live in Florida where we can get cool nights and hot days so there was some variability. I then tested the samples on day 5, day 10, and day 19…Below is the results. (CHARTS 1-4)
CHART 1
CHART 2
CHART 3
CHART 4
From these results a question arose as to whether or not the composition of the container could have an impact on the results. A second experiment was conducted using containers made of different materials…Glass, Polypropylene, High Density Polyethylene, Polyethylene, PET, Poly-carbonate and Polystyrene. I chose to test only the PO4 because it had the biggest change over time as well as this was the parameter that posed the question. In addition the measurements were taken using two different Hanna Testing Meters... the HI-736 ULR Phosphorous Meter and the HI-774 ULR Phosphate Meter. The tests were replicated at least 2 times for confirmation some were done as many as 4 times. There are a lot of charts (23) resulting from this set of experiments that I won’t include here, but if you are interested I can send you the charts.
The initial findings indicated that there were measurable differences between the different containers, with glass and PET giving the best results. Then @Dan_P raised the question could the Surface Area to Volume ratio (SA/V) be a factor given that the containers were of different sizes. Sure enough his hunch was correct. The containers with the highest SA/V ratio had the worst results. This initiated a third round of experiments using containers with equivalent SA/V ratios. I chose to use the actual sample tubes provided by the ICP vendor that are made from Polypropylene (PP) and glass sample tubes with the same volume and dimensions thus giving the same SA/V ratio. These tests were also done in duplicate. The results are below. (CHARTS 5 & 6)
CHART 5
CHART 6
At this point a question was posed by as to whether this was due to Bio-Activity or purely a chemical reaction. I had remembered reading that one of the ICP vendors had tested the question of storage stability using a HACH standard solution and testing it over several days. They reported that the measurement was stable over the storage period. I decided to replicate this experiment.
I made up a PO4 standard at .06 ppm using salt water at 35ppt made with RODI water and reagent grade Sodium Chloride and a HACH Standard. The thinking was that this should represent a Bio-Inactive sample. The same test protocol that was used for the tank sample was used. I chose two different containers, glass and a HDPE, the results are below: (CHARTS 7 & 8)
CHART 7
CHART 8
The test results indicated that although there was a change over the storage period the results were within testing error and did not show the significant change found in the tank samples that were Bio-Active.
After reviewing the test procedure of the HACH Standard experiment I had some question as to whether the containers were completely free of any Bio material. I did a repeat of the test after sterilizing both containers. Results are below: (CHARTS 9 & 10)
CHART 9
CHART 10
This would be an indicator that the instability of the PO4 is related to the Bio-Activity of the tank samples.
One hypothesis is that the Bio-Active samples from the tank are forming a thin bio-film on the sides of the containers that over time adsorb the organic phosphate their by reducing the amount of PO4 in the water being tested. This is only one of several possible explanations yet to be explored.
An additional experiment was set up to see it ‘Pre-Conditioning” the container might be helpful. The thought was by doing this the film would be formed and the tubes would be “conditioned” and the next sample placed in the tube would not undergo the “film forming” because the film was already there.
The PP sample tubes were filled with tank water and allowed to set for 12 days. An initial measurement was taken (Day 0) and then a second measurement on day 12. The tubes were then re-filled with a second batch of tank water without rinsing out the tubes. The second batch was then tested according to the standard test protocol. (Store samples and test on day 0, 3 & 6 and test with the 2 Hanna Checkers). The results are below: (Charts 11 & 12)
CHART 11
CHART 12
It also could be that the hypotheses is incorrect and there is some other mechanism causing this…Bacteria interaction…some type of chelation process….and who know what else….Most of this is way above my pay grade, but many of you out there in R2R Family have the skills, knowledge and ability to contribute ideas.
SOME INITIAL CONCLUSIONS:
- Samples stored in high SA/V containers (Sample Tubes) are not stable over time with respect to PO4 measurement when tested with Hanna Checkers.
- The problem is associated with tank water (Bio-Active) and does not show up with “sterile” samples made with PO4 standards.
- The type of container does not appear to make a difference. The SA/V Ratio appears to be the controlling factor.
SOME INITIAL SPECULATIONS:
- This could be a possible explanation of why a number of Reefers get significantly different results from there testing protocol and ICP results when testing PO4.
- There could be other test parameters that are affected by this issue.
- A different sampling protocol may be necessary to insure better ICP test results related to some test perimeters.
Well that is about the long and the short of it at this point. The work is continuing and I will provide updates as they are completed. There are so many “Rabbit Holes” to this project! I have several questions yet to be explored….
- Does the initial PO4 level change the outcome?
- Are there other parameters to be concerned about? Evidence says Iodine might be an issue…
- What is the effect of temperature; would keeping the sample cold reduce the problem?
- Is there a critical Surface Area to Volume Ration where the problem is reduce or eliminated and what is it?
- Why are there some cases where the ICP results closely match the results obtained by the Hanna Checkers and yet there are other times they are vastly different (2-3 times)?
Thanks again to @Dan_P and @taricha for their help and guidance in this project...What an awesome family this R2R group is!!
One thing is for sure BRS is happy about this project. With over 200 individual tests and more to come their sales of Checker Reagent is benefiting greatly!