I got the magnesium checker about a week ago.
It's very easy to use, comparable to the Alkalinity checker in terms of ease of use. There's only one extra liquid reagent to mix with the unreacted sample water, when concocting the "blank" cuvette for C1.
When I first got it, I first checked its accuracy against a one year old bottle of ATI reference 1000 ml.
The bottle said 1313 ppm magnesium, the tester said 1380 ppm. My bad, old bottle and nearly used up.
Today, I checked its accuracy again against a new bottle of ATI reference that I've just got in a few hours ago.
Again, the ICP-tested bottle says 1313 ppm magnesium, but the test this time says 1345 ppm. Closer.
It seems to me that - in terms of accuracy - this is comparable to a titration test kit, like the Salifert one, just much easier, quicker, and with a digital read-out.
They both seem to have that plus minus 30 ppm range of uncertainty.
Perhaps, if I found a better way to measure the quantity of liquid reagents and water (instead of the provided syringes), the accuracy could be better.
Finally, in this test there's no powder reagent to get out of hellish foil pouches. That alone is a reason to rejoice!
It's very easy to use, comparable to the Alkalinity checker in terms of ease of use. There's only one extra liquid reagent to mix with the unreacted sample water, when concocting the "blank" cuvette for C1.
When I first got it, I first checked its accuracy against a one year old bottle of ATI reference 1000 ml.
The bottle said 1313 ppm magnesium, the tester said 1380 ppm. My bad, old bottle and nearly used up.
Today, I checked its accuracy again against a new bottle of ATI reference that I've just got in a few hours ago.
Again, the ICP-tested bottle says 1313 ppm magnesium, but the test this time says 1345 ppm. Closer.
It seems to me that - in terms of accuracy - this is comparable to a titration test kit, like the Salifert one, just much easier, quicker, and with a digital read-out.
They both seem to have that plus minus 30 ppm range of uncertainty.
Perhaps, if I found a better way to measure the quantity of liquid reagents and water (instead of the provided syringes), the accuracy could be better.
Finally, in this test there's no powder reagent to get out of hellish foil pouches. That alone is a reason to rejoice!