Good to know about the back pressure and where the fluid has to be. Thank youThat is not calibrating. That is setting the Trident to the other testers. In order to calibrate correctly, the calibration solution must be at the same level as the end of the sample tube when in normal operation. The tube must also be ran in the same configuration as normal. The back pressure on the tube makes a difference.