Alkatronic Issues

Ecotech Marine

m.kristoff

Active Member
Review score
+3 /0 /-0
View Badges
Joined
Sep 17, 2021
Messages
458
Reaction score
218
Review score
+3 /0 /-0
Location
cleveland
Rating - 100%
1   0   0
I setup My alktronic this past Sunday.
My 1st reading came back with 9.09. I crossed checked to with my Hanna which was 9.0. I thought it can't get any better than that! It went downhill from there!

Next test came back with 9.77. I did a manual retest which came back 9.74. I turned off my Kalkawasser dosing just to rule out a spike. The next test came back with 9.71.
I emailed Foucstronic and they said give it a week to settle in- Not sure I am on board with that but why not.

For the rest of the day I seen consistent range between 9.74 trending downward to 9.54. It held 9.54 for 3 test so I figured it was indeed settling in. My Hanna came back with 9.1.

I decided to rule out calibration errors, I recalibrated the pumps and mixed up new reagent using a scale.
Next test came back with 9.4.
the following test came back with 5.12- not sure why it didn't automatically retest
it tested 3 hours later and it was 9.60
as of the last test it is 9.77

I believe my dkh is really about 9.0ish based on the hanna and I had a trident and its last test before I took it offline was 8.75.

Should I consider a reset and start over or let it run and see if it settles in?

dkh-21st.jpg dkh-22nd.jpg
 
Last edited:
www.dinkinsaquaticgardens.com
OP
OP
m.kristoff

m.kristoff

Active Member
Review score
+3 /0 /-0
View Badges
Joined
Sep 17, 2021
Messages
458
Reaction score
218
Review score
+3 /0 /-0
Location
cleveland
Rating - 100%
1   0   0
If they suggested letting it run for a week I'd probably try that first. Even the Trident needs a few days to settle in and be recalibrated.
I agree and am doing that. it is constant even though it is reading it high. just making sure I am ruling out any errors I may have caused with setup
 
Nutramar Foods

blaxsun

10K Club member
Review score
+0 /0 /-0
View Badges
Joined
Dec 15, 2020
Messages
21,340
Reaction score
25,794
Review score
+0 /0 /-0
Location
The Abyss
Rating - 0%
0   0   0
I agree and am doing that. it is constant even though it is reading it high. just making sure I am ruling out any errors I may have caused with setup
I took me a month or two playing around with my Trident to get it to deliver consistent results, too. I still test to "keep it honest" - but I'm not manually testing anywhere near as much as I used to. How's the pH in terms of accuracy?
 

Oldreefer44

Valuable Member
Review score
+0 /0 /-0
View Badges
Joined
May 31, 2016
Messages
1,300
Reaction score
1,749
Review score
+0 /0 /-0
Location
Machias Washington
Rating - 0%
0   0   0
Mine always runs about .6 above my Hanna. So being more concerned with consistency than an actual number, as long as it is close to my target, it really doesn't bother me. It has been much more consistent than the Trident I had.
 
OP
OP
m.kristoff

m.kristoff

Active Member
Review score
+3 /0 /-0
View Badges
Joined
Sep 17, 2021
Messages
458
Reaction score
218
Review score
+3 /0 /-0
Location
cleveland
Rating - 100%
1   0   0
Mine always runs about .6 above my Hanna. So being more concerned with consistency than an actual number, as long as it is close to my target, it really doesn't bother me. It has been much more consistent than the Trident I had.
I got rid of my trident because of inconsistency. My numbers are very high compared to my hanna.. it just came back with a 12.5!
 
CLICK TO VIEW
OP
OP
m.kristoff

m.kristoff

Active Member
Review score
+3 /0 /-0
View Badges
Joined
Sep 17, 2021
Messages
458
Reaction score
218
Review score
+3 /0 /-0
Location
cleveland
Rating - 100%
1   0   0
How's the pH in terms of accuracy?
PH was spot on. verified with Hanna handheld. I also recalibrated the PH prob just to rule out any mistakes there as well. currently my hanna handheld is 8.1 (reciently calibrated) now after the probe calibration, its off. its reading 7.78

it just ran another test and its 11.64 so it is actually worse now
 

MGarrison

Active Member
Review score
+1 /0 /-0
View Badges
Joined
Jul 23, 2020
Messages
173
Reaction score
132
Review score
+1 /0 /-0
Location
Columbus
Rating - 0%
0   0   0
1) it does take a few days for it to settle in. Even when changing out the pump hoses. you will want to let it run a few days then recalibrate again.
2) Where you are drawing tank water from or Reagent container, could there be a possibility of it pulling air bubbles?
3) check the height of your ph prob. I had to add a rubberband to raise it a bit to clear the magnetic pill.
4) Pipe from pump B must be in contact with the bottom. Very important!!
5) Pipe from pump A must be around 2 mm above bottom. Very important!!
 
OP
OP
m.kristoff

m.kristoff

Active Member
Review score
+3 /0 /-0
View Badges
Joined
Sep 17, 2021
Messages
458
Reaction score
218
Review score
+3 /0 /-0
Location
cleveland
Rating - 100%
1   0   0
1) it does take a few days for it to settle in. Even when changing out the pump hoses. you will want to let it run a few days then recalibrate again.
2) Where you are drawing tank water from or Reagent container, could there be a possibility of it pulling air bubbles?
3) check the height of your ph prob. I had to add a rubberband to raise it a bit to clear the magnetic pill.
4) Pipe from pump B must be in contact with the bottom. Very important!!
5) Pipe from pump A must be around 2 mm above bottom. Very important!!
please see photos. everything is set as you suggested. I have moved the intake to 3 different places just to rule out air bubbles. it does have the filter on it. It has been running since sunday and is getting worse.
I did recalibrate probe using Milwaukee 7.01 and Pinpoint 4.000 fluid. I wouldn't think that would be an issue
 

Attachments

  • 1.jpg
    1.jpg
    93.5 KB · Views: 38
  • 2.jpg
    2.jpg
    77.9 KB · Views: 42
OP
OP
m.kristoff

m.kristoff

Active Member
Review score
+3 /0 /-0
View Badges
Joined
Sep 17, 2021
Messages
458
Reaction score
218
Review score
+3 /0 /-0
Location
cleveland
Rating - 100%
1   0   0
the intake is in a different chamber . The only thing I dose is kalkawasser and I have turned that off to validate that is was not interfering and it is not.
I just moved the unit to its final spot. recalibrated the pumps and got an expected reading of 9.09 and I verified with my Hanna at 9.1 . Next tests are now really low. cant figure it out.
 

Attachments

  • Screenshot 2022-03-22 205318.jpg
    Screenshot 2022-03-22 205318.jpg
    39 KB · Views: 39
www.dinkinsaquaticgardens.com
OP
OP
m.kristoff

m.kristoff

Active Member
Review score
+3 /0 /-0
View Badges
Joined
Sep 17, 2021
Messages
458
Reaction score
218
Review score
+3 /0 /-0
Location
cleveland
Rating - 100%
1   0   0
Black dots mean washout.
Follow the bad precision section of the manual.
I did and validated that there is no air in the line.

the 9.09 value is after a restart if that makes any difference.

the machine is proving to have extreme swings. from a high of 12.7 down to 8.06. not sure what else I can do
Washout mode is turned off
 

Robinson

Active Member
Review score
+0 /0 /-0
View Badges
Joined
May 9, 2016
Messages
499
Reaction score
1,034
Review score
+0 /0 /-0
Location
Miramar
Rating - 0%
0   0   0
I worked with Eric @ focustronic and he had me test pump A and C. they were both off. Recalibrated and the readings are back in the mid 9's now. I think it probably needs to settle in View attachment 2601319

That looks fine for me, what I compare is the previous day test at the same time with the current day. Also remember the Hanna have accuracy of +/- 0.3 DKH, so when you compare to that equipment you need to take that in consideration. Like someone already mentioned what is important is the consistency than the actual number.
 
Nutramar Foods
OP
OP
m.kristoff

m.kristoff

Active Member
Review score
+3 /0 /-0
View Badges
Joined
Sep 17, 2021
Messages
458
Reaction score
218
Review score
+3 /0 /-0
Location
cleveland
Rating - 100%
1   0   0
That looks fine for me, what I compare is the previous day test at the same time with the current day. Also remember the Hanna have accuracy of +/- 0.3 DKH, so when you compare to that equipment you need to take that in consideration. Like someone already mentioned what is important is the consistency than the actual number.
Agreed that consistency is key. still need that baseline! I think ill let it run for a week and watch the trends. Ill recalibrate in about 7 days or so.
 
AI Nero 5

Polyp polynomial: How many heads do you start with when buying zoas?

  • One head is enough to get started.

    Votes: 27 10.6%
  • 2 to 4 heads.

    Votes: 145 57.1%
  • 5 heads or more.

    Votes: 65 25.6%
  • Full colony.

    Votes: 10 3.9%
  • Other.

    Votes: 7 2.8%
Back
Top