Might be a stupid question about hanna checkers, but here it goes.

Was running a test on the nitrate hr checker and got a result of 46 ppm. Seeing as this test was a few hours after a roughly 20% water change, I assumed it was a bad test. Pulled the vile and noticed small bubbles throughout, figured it caused a bad result.

Got a clean vile, zeroed the test again, retested the same original sample, new result was 22.8 ppm. Performed the same procedure and got 23.0 ppm.

Can you use the same reagent multiple times as I did above and get accurate results?
 

KrisReef

10K Club member
View Badges
Joined
May 15, 2018
Messages
15,227
Reaction score
31,279
Location
ADX Florence
Rating - 0%
0   0   0
I have found that my Checkers will give repeatable results when I am careful to use proper technique each time.
When I get weird results I retest to see if it was me or the tank that is off.
 
Upvote 0

KrisReef

10K Club member
View Badges
Joined
May 15, 2018
Messages
15,227
Reaction score
31,279
Location
ADX Florence
Rating - 0%
0   0   0
The exception to this may be dKH if the reagent is getting old. It is known to fade after being opened and used for a long period. I open a new reagent to see what is up if the bottle is older and the results are unexpected.

Stability is expected so deviation requires validation.
 
Upvote 0

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
72,100
Reaction score
69,741
Location
Massachusetts, United States
Rating - 0%
0   0   0
Opening up 1 packet of reagent, placing that into the sample container, and running multiple tests on that sample, say 3 times to get an average .

I am confused. I've never used this kit, but if we are talking about the HI 781, in looking at the directions, I do not see a step where there is "left over" reagent from one test that you might be using for a second test later.

 
Upvote 0
OP
OP
KochergenJ

KochergenJ

Community Member
View Badges
Joined
Jan 29, 2024
Messages
81
Reaction score
76
Location
Fresno
Rating - 0%
0   0   0
I am confused. I've never used this kit, but if we are talking about the HI 781, in looking at the directions, I do not see a step where there is "left over" reagent from one test that you might be using for a second test later.


Sorry, my bad on the explanation.

So I put 10 ml into a vile. Place the entire packet of reagent into the vile. Shake for 2 minutes, I know have my test sample.

With a second vile, I go pull 10 ml of tank water, no reagent in this vile, this is now my "zeroing" solution for the test.

Can you run the test sample multiple times or do you need to empty it out after each test and open a new packet of reagent?
 
Upvote 0

Reef Wizard

Octopus Enthusiast
View Badges
Joined
Apr 1, 2022
Messages
1,224
Reaction score
3,178
Location
Middle TN
Rating - 0%
0   0   0
I am confused. I've never used this kit, but if we are talking about the HI 781, in looking at the directions, I do not see a step where there is "left over" reagent from one test that you might be using for a second test later.

I think he maybe is referring to using the same cuvette and reacted sample again after testing it once. Instead of using another unreacted sample and new reagent. pardon my chemistry french lol not sure of all the terms but I try
 
Upvote 0

Uncle99

10K Club member
View Badges
Joined
Jul 22, 2018
Messages
10,506
Reaction score
15,974
Location
Province of Ontario
Rating - 0%
0   0   0
Opening up 1 packet of reagent, placing that into the sample container, and running multiple tests on that sample, say 3 times to get an average .
You can do that but nitrate will darken over minutes which may lead to incorrect assumptions.

At 2 minutes, the shake time, it will have coloured up, in the next 7 minutes, the wait time, the colour will intensify, so if you retest at say 10 mins, the result may be higher. There would be a point where the colour has fully developed.

If a test comes back largely outside your “expected” result, just rerun the full test again.

There are many ways tests come out incorrect.
 
Upvote 0
OP
OP
KochergenJ

KochergenJ

Community Member
View Badges
Joined
Jan 29, 2024
Messages
81
Reaction score
76
Location
Fresno
Rating - 0%
0   0   0
You can do that but nitrate will darken over minutes which may lead to incorrect assumptions.

If a test comes back largely outside your “expected” result, just rerun the the test.
You can do that but nitrate will darken over minutes which may lead to incorrect assumptions.

If a test comes back largely outside your “expected” result, just rerun the the test.

That is what I was unsure of. If the sample continues to darken over time or not. To run 3 tests would be roughly 21 minutes as the tester takes 7 minutes to perform. Wasn't sure how the "sample" would react over that duration.
 
Upvote 0

Uncle99

10K Club member
View Badges
Joined
Jul 22, 2018
Messages
10,506
Reaction score
15,974
Location
Province of Ontario
Rating - 0%
0   0   0
That is what I was unsure of. If the sample continues to darken over time or not. To run 3 tests would be roughly 21 minutes as the tester takes 7 minutes to perform. Wasn't sure how the "sample" would react over that duration.
After 7 minutes we would expect colour to be mostly developed, so if retested at 14 minutes, the results should be close.

All your really doing is asking the Hanna to recheck the colour.

The colour will only darken so far regardless of time, so if I did a second check, I’d do it as close to the 7 minutes as possible.
 
Upvote 0

DanyL

Valuable Member
View Badges
Joined
Jan 13, 2023
Messages
1,477
Reaction score
1,738
Location
Middle East
Rating - 0%
0   0   0
I do this from time to time with the Phosphate ULR test.

However, instead of “press and hold” to start the timer, I do a regular press which samples immediately, and it needs to be done rather quickly.

This only applicable when there was a problem with the vial itself, like a missed air bubble or an accidental smudge you’ve only noticed after the fact and caused the result to be invalid.
 
Upvote 0

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
72,100
Reaction score
69,741
Location
Massachusetts, United States
Rating - 0%
0   0   0
I'd be wary of repeating the absorbance measurement at longer and longer times to try to get a better reading as many of the tests we use are not stable. Some companies explicitly say what window of time is acceptable and what is not.

The Hach nitrate kit says, for example:

"Allow at least 10 minutes, but not more than 20 minutes, before..."
 
Upvote 0
OP
OP
KochergenJ

KochergenJ

Community Member
View Badges
Joined
Jan 29, 2024
Messages
81
Reaction score
76
Location
Fresno
Rating - 0%
0   0   0
I do this from time to time with the Phosphate ULR test.

However, instead of “press and hold” to start the timer, I do a regular press which samples immediately, and it needs to be done rather quickly.

This only applicable when there was a problem with the vial itself, like a missed air bubble or an accidental smudge you’ve only noticed after the fact and caused the result to be invalid.

I had missed a lot of air bubbles and realized after the test results. Got rid of those bubbles, ran it twice after and results were within .02 ppm.
 
Upvote 0

DanyL

Valuable Member
View Badges
Joined
Jan 13, 2023
Messages
1,477
Reaction score
1,738
Location
Middle East
Rating - 0%
0   0   0
I had missed a lot of air bubbles and realized after the test results. Got rid of those bubbles, ran it twice after and results were within .02 ppm.
If the vials were shaken between the tests it’ll likely affect the results as well, wether it being the reaction itself or by indissolved powder floating around inside the sample.

For air bubbles, when I miss them it’s usually just one or two, and slightly rotating the vial in a way that the sensor won’t be abstracted by it is enough.

For smudges I do the same when possible, otherwise I carfully clean it without disturbing the sample in the process.

I’ve done this quite a few times, and when done correctly it aligns perfectly with my trends.
 
Upvote 0

Reefer Matt

Reef Cave Dweller
View Badges
Joined
May 15, 2021
Messages
6,977
Reaction score
31,412
Location
Michigan
Rating - 0%
0   0   0
Yes you can do that so long as it is not hours in between tests, as the results may not be accurate. In fact, if you miss the countdown and the checker turns off, you usually don’t need to retest with new reagent. Just fill another cuvette with tank water, zero it out (C1), then put the original colored sample in for C2. Press the button instead of holding to get an immediate result.

The length of time the reagent can sit in the cuvette is debatable (I will ask Hanna), but if you just missed the timer by a few minutes or so, this will save you some reagent. You can also retest with the same filled cuvettes to get an average, but the results should be very close.

Here’s a video so you don’t have to take my word for it, but I have also tested this myself.

 
Upvote 0