(Almost) Everything You Need to Know About Apex Salinity Calibration and TC Factor

eag

Active Member
View Badges
Joined
Apr 3, 2020
Messages
323
Reaction score
412
Location
Bay Area, California
Rating - 0%
0   0   0
I was trying to calibrate my new salinity probe a few weeks ago, and ran into all sorts of problems. Calibrations that were way off the mark, big fluctuations in readings that couldn't possibly be correct (and were verifiably not), etc. After searching around the internet, it was clear that a lot of folks have problems with this sensor. Some have written it off completely, and others insist that such problems are the result of calibration error. Information on correct calibration and configuration was present but scattered. Some of it was incomplete or incorrect. After having read many forum threads, spoken to Apex customer service, and experimented with different approaches, I am here to tell you that the sensor does work reliably when correctly configured and calibrated. This post is intended to capture everything I've learned in a single place, hopefully making this easier for someone else in the future.

Apex Configurables
There are only two configurable aspects of the Apex salinity sensor that can affect the reading. The first is, of course, the calibration point. The second is the temperature compensation factor (or TC factor). The salinity probe's readings are significantly affected by temperature - the TC factor setting allows you to compensate for this if you also have a temperature probe configured. The rest of this post captures the concepts and steps necessary to configure these inputs correctly.

TC Factor and Correctness
The TC factor configurable can be found on the advanced tab of the conductivity probe configuration page. It is a percentage, set manually - apparently different tanks land on different values. The percentage controls exactly how much compensation should be applied (i.e. x% compensation per degree).

One way to see if your TC factor is correct is by looking at your salinity graph with your temperature readings overlaid. If the TC factor needs to be adjusted, you will be able to see that the salinity readings track fluctuations in temperature. If the TC factor is correct(-ish), this correlation will not be present.

Here is what this looks like when TC is disabled (i.e. 0%):
Screenshot 2020-05-10 at 15.17.10.png


Here is the same system with TC configured:
Screenshot 2020-05-10 at 15.18.09.png


You can see from the first screenshot a clear correlation between temperature and salinity. Compare to the second screenshot with TC applied, and the salinity is relatively stable. I'm still not quite sure about those little bumps though. I continue to play around fine tuning the TC factor to see how they react (if at all - it's possible that the fluctuation is related to some funny business with my overflow).

Configuring the TC Factor
The TC factor defaults to 0. This is confusing for new Apex owners, as the salinity probe is all but worthless without applying temperature compensation. After speaking with customer service, and reading through some documentation, it appears that the default is 0 because setting it to anything other than zero causes problems in systems that don't have a temperature probe installed (the documentation includes a large WARNING section about this). Since Neptune cannot know or guarantee that any given system will have its temperature probe installed, the default is 0 to be safe.

If you have a temperature probe installed, you will definitely want to adjust the default TC factor. It is important to note that the temperature probe must be plugged into the same module as the salinity probe. In other words, if your salinity probe is plugged into your base unit, then you must also have a temperature probe plugged into the base unit. Similarly, if your salinity probe is plugged into a PM2 salinity module, then the temperature probe must be plugged into the same PM2.

TC Factor Value Range
It may take some experimentation to find the correct TC factor for your system. The Apex documentation states that seawater normally has a TC factor between 2.1% and 2.3%, thus recommending 2.2% as a good starting point. By comparison, a Neptune representative told me that in their experience, correct values are often found in the 1.9% to 2.1% range. Your mileage may vary.

The Apex documentation states that the salinity probe must be re-calibrated when changing the TC factor. According to a Neptune representative with whom I spoke, this is only partially true. When TC factor is non-zero and the salinity probe is configured, the current water temperature is stored inside the salinity calibration point that Apex records. This temperature is used to apply temperature compensation going forward. If the TC factor is zero when the salinity probe is configured, then the current water temperature is not stored, making future temperature compensation impossible. Therefor, in order for TC factor to take effect, it must be configured with a non-zero value at the time that the salinity probe is calibrated.

Fear not, however, in your search for the correct TC factor value. The Neptune representative informed me that once the salinity probe calibration point includes temperature data, the TC factor value can be changed at will without requiring re-calibration. Changes to the TC factor will be applied to all salinity readings going forward - readings that were taken in the past will not change.

How to Configure TC Factor for the First Time
  1. Ensure that a temperature probe is plugged into the same module as your salinity probe
  2. Navigate to the `Advanced` tab of the salinity probe configuration
  3. Change the TC factor value to any non-zero positive number (2.1 is a good starting point)
  4. Click "Update Apex" in the top right corner to upload the new configuration
  5. Perform a salinity calibration
  6. Salinity readings going forward will now have temperature compensation applied accordingly
Calibrating Salinity
There are a number of factors that require attention when calibrating the salinity probe. If any one of them are ignored, the resulting readings will be incorrect. Some of these are described in the documentation, and some are not.

We have learned that salinity readings are affected by temperature. As a result, temperature plays a very important part in the calibration process. When temperature compensation is disabled, and the Apex salinity calibration solution is used (the grey pouch, 53,000 microsiemens), the temperature during calibration is assumed to be 77.0 degrees fahrenheit. This is rarely actually the case, and folks calibrating their salinity probes who aren't aware of this detail will almost certainly experience inaccurate readings.

If however, you have a temperature probe, you do not need to warm the calibration solution to precisely 77.0 degrees. Apex will instead calibrate the salinity using the temperature reading provided by the probe. Note that this still means that the solution must be at precisely the same temperature as the probe! To accomplish this, you can simply float the calibration solution pouch in water near the temperature probe location for 15 minutes. Note that the calibration solution temperature can change rapidly - for instance, holding the pouch in your hand while performing the calibration will affect the accuracy. Be sure to keep the pouch in the water during the entire calibration process to ensure that the solution temperature doesn't deviate.

How to Calibrate Salinity
  1. Ensure that your TC factor has already been configured with a non-zero value
  2. Float the calibration solution pouch near your temperature probe for 15 minutes
  3. Navigate to the `Basic` tab of the salinity probe configuration
  4. Click `Automatic Calibration`
  5. When asked to dry the salinity probe, remove it from the water and rinse with RO/DI. Then gently shake the probe, and pat the bottom dry with a paper towel
  6. When asked to insert the salinity probe into the calibration solution, open the pouch and insert the probe. Be sure to keep the pouch submerged in water during this process so the temperature remains stable
  7. Gently shake the probe in the calibration solution to remove any bubbles that might be trapped. This step is VERY important - shake for a little longer than you might otherwise
  8. Wait for the readings to stabilize, and complete the calibration. You're all done!
Note that the manual calibration process might work here as well, though I've never done that and so have no idea what it does or doesn't involve :)

Finding More Information
I spent several days searching for information on this topic. This post is, as far as I know, the most complete source of information on Apex salinity calibration and temperature compensation currently available. As a new Apex owner, I was shocked to find that none of this information was included in any of the documentation I could find. When asking Neptune customer support about this, they referred me to the PM2 documentation. There is definitely more information there than the "regular" Apex docs, however it is still not complete and does not fully apply to the use of the base module for salinity readings.

If you find that any of the information in this post is inaccurate, or if you have information that is not included here but should be, please post a comment on this thread!!

Summary
I have found the Apex salinity sensor to be relatively accurate when correctly configured. Of particular importance is configuring temperature compensation, ensuring stable temperature conditions during calibration, and ensuring the correct order of operations. I hope that others who may be struggling with their Apex salinity sensor are able to use this thread to fix things up. Happy reefing!
 

SuncrestReef

That Apex guy
View Badges
Joined
Jan 18, 2018
Messages
4,214
Reaction score
9,251
Location
Oregon
Rating - 0%
0   0   0
Excellent write-up. Like you, I have not seen this much detail on TC in a single place.

One thing I will add regarding the salinity probe, and that I have shared many times on this and other forums, is eliminating the issue of trapped bubbles causing problems. I keep my probe mounted upside down in my sump so no bubbles can accumulate and get trapped inside the probe. I also calibrate my probe upside down with the calibration solution inside a probe storage bottle, so I can invert it and shake it allowing the bubbles to float up out of the tube. If you don't have a probe bottle, you can achieve the same result by using a small plastic bag containing the calibration solution, then use a rubber band to hold it onto the probe while tipping it upside down.

1589156135546.png

 

((FORDTECH))

2500 Club Member
View Badges
Joined
Jan 23, 2020
Messages
4,838
Reaction score
4,279
Location
Chicago
Rating - 0%
0   0   0
Great write up I will be Messing with that TC a bit as I just set it to the 2.2 awhile ago cause that was recommended and seemed to get me in ball park. Also like to say I do the exact same as suncreast does with mounting probe upside down to eliminate bubbles getting trapped and works like a charm.
 

glb

7500 Club Member
View Badges
Joined
May 12, 2013
Messages
8,141
Reaction score
3,367
Location
Miami
Rating - 0%
0   0   0
So I followed these directions to a “t”. Set TC to 2.2 and calibrated. Right after calibrating correctly, my conductivity in the tank is reading 36.5. I used to have my temp probe on the ph module but just moved it to salinity because the salinity readings went up and down with the temp. I know the tank is between 34-35. Any thoughts on why I’m getting the high reading? Thanks so much for posting this btw. It’s so helpful!
 

dwest

2500 Club Member
View Badges
Joined
Feb 27, 2018
Messages
4,605
Reaction score
9,586
Location
Northern KY
Rating - 0%
0   0   0
Awesome! The temperature correlation always bothered me some. I had no idea we could compensate.
 
OP
OP
eag

eag

Active Member
View Badges
Joined
Apr 3, 2020
Messages
323
Reaction score
412
Location
Bay Area, California
Rating - 0%
0   0   0
So I followed these directions to a “t”. Set TC to 2.2 and calibrated. Right after calibrating correctly, my conductivity in the tank is reading 36.5. I used to have my temp probe on the ph module but just moved it to salinity because the salinity readings went up and down with the temp. I know the tank is between 34-35. Any thoughts on why I’m getting the high reading?

Does the TC appear to be working?

I had a similar experience after getting my probe calibrated. It turned out that my refractometer was incorrect. I had been calibrating with RO/DI, but that was not sufficient. I ordered a bottle of 1.026 calibration fluid, which read at ~1.0235 after RO/DI calibration :oops:.
 
OP
OP
eag

eag

Active Member
View Badges
Joined
Apr 3, 2020
Messages
323
Reaction score
412
Location
Bay Area, California
Rating - 0%
0   0   0
@glb the video that @powers2001 shared indicates that cable routing could affect readings too, though I did not see an issue there in practice. Assuming the equipment you're using to assert you 34-35ppt salinity is reading accurately, perhaps check to make sure that your probe cable is not routed alongside any higher voltage power cables?
 

thermoJoe

Active Member
View Badges
Joined
Jun 9, 2013
Messages
142
Reaction score
240
Location
Marshfield MA
Rating - 0%
0   0   0
Great idea! This hasn't been a problem in my tank yet, but I've heard of others encountering this.
Nice write-up. I'm not sure if the probe is truly designed to be submerged, so you might find that submerging it reduces its life expectancy.

The problem I have is that brittle stars climb into it every now and then causing bogus readings.
 

Flameback Pair

started reefing 1999
View Badges
Joined
May 2, 2013
Messages
6,005
Reaction score
5,690
Location
LIVING WELLS MONTANA
Rating - 0%
0   0   0
Nice write-up. I'm not sure if the probe is truly designed to be submerged, so you might find that submerging it reduces its life expectancy.

The problem I have is that brittle stars climb into it every now and then causing bogus readings.
@eag I would like to know this too whether the salinity probe can be completely submerged. I’ve read all Neptune probes with the blue rubber caps are submersible.
 
OP
OP
eag

eag

Active Member
View Badges
Joined
Apr 3, 2020
Messages
323
Reaction score
412
Location
Bay Area, California
Rating - 0%
0   0   0
@eag I would like to know this too whether the salinity probe can be completely submerged. I’ve read all Neptune probes with the blue rubber caps are submersible.

The documentation has this to say:
Probes can be fully or partially submerged as they are all fully sealed against water, however the cable insulation can become brittle with constant immersion in salt water. Also, it is important to ensure that any of the bulb type probes (pH, ORP) stay wet at all times as they will be damaged if they dry out.

My interpretation of that statement is: it will work, it may reduce the life expectancy (as noted by @thermoJoe), and that you should be careful about letting lower water levels expose the bulb if mounted upside down. If it were me, I'd probably also watch out for the bend radius of any portion of the cable that is regularly submerged.
 

Miller535

Valuable Member
View Badges
Joined
Dec 11, 2019
Messages
2,203
Reaction score
1,939
Location
Pittsburgh, Pennsylvania
Rating - 0%
0   0   0
Excellent write-up. Like you, I have not seen this much detail on TC in a single place.

One thing I will add regarding the salinity probe, and that I have shared many times on this and other forums, is eliminating the issue of trapped bubbles causing problems. I keep my probe mounted upside down in my sump so no bubbles can accumulate and get trapped inside the probe. I also calibrate my probe upside down with the calibration solution inside a probe storage bottle, so I can invert it and shake it allowing the bubbles to float up out of the tube. If you don't have a probe bottle, you can achieve the same result by using a small plastic bag containing the calibration solution, then use a rubber band to hold it onto the probe while tipping it upside down.

1589156135546.png


How do you do this with keeping the wire and boot on the top of the probe out of the water? As Neptune says this will cause damage to the rubber on the top of the probe?

I turn mine on a 45 degree angle so that the wire is still out of the water, but the bottom of the sensor faces away from the direction the water is flowing through the sump. If that makes any sense. lol
 

Miller535

Valuable Member
View Badges
Joined
Dec 11, 2019
Messages
2,203
Reaction score
1,939
Location
Pittsburgh, Pennsylvania
Rating - 0%
0   0   0
I am one of the people that have long argued with those who have called the Apex conductivity probe "junk", and told them that it was either calibration or micro bubbles causing them an issue.

My salinity probe sits VERY still on 34.5, while I get a reading of 35 with my refractometer. That is 1.0260 compared to 1.0264. I call that really accurate.

The only problem I have ever had with the probe in the year and a half that I have been using it was about 3 weeks ago it get dropping lower and lower to like 30. something. I decided hey I haven't cleaned or calibrated this thing in about 6 months. So I soaked it in a 1:1 Vinegar to water solution for about 15 mins. Scrubbed it gently with a soft brush. Recalibrated it, and it's rock solid again.
 
OP
OP
eag

eag

Active Member
View Badges
Joined
Apr 3, 2020
Messages
323
Reaction score
412
Location
Bay Area, California
Rating - 0%
0   0   0
The only problem I have ever had with the probe in the year and a half that I have been using it was about 3 weeks ago it get dropping lower and lower to like 30. something. I decided hey I haven't cleaned or calibrated this thing in about 6 months. So I soaked it in a 1:1 Vinegar to water solution for about 15 mins. Scrubbed it gently with a soft brush. Recalibrated it, and it's rock solid again.

Oh, this is good to know... if there are other reports of this, perhaps it is worth adding to the original post. Interesting that it started drifting "all of a sudden" ... Maybe precipitation or some other event that led to the need to clean it?
 

Miller535

Valuable Member
View Badges
Joined
Dec 11, 2019
Messages
2,203
Reaction score
1,939
Location
Pittsburgh, Pennsylvania
Rating - 0%
0   0   0
Oh, this is good to know... if there are other reports of this, perhaps it is worth adding to the original post. Interesting that it started drifting "all of a sudden" ... Perhaps precipitation or some other event that led to the need to clean it?

Yeah, I saw several threads on this. The probes do get precipitant, and sometimes they even get those little tube worm looking things in them. Although on the threads some claimed actual probe failure or pm2/brain failure. So if cleaning it did not fix the problem, it looks like it's either the probe or module/brain the probe is connected to.
 
OP
OP
eag

eag

Active Member
View Badges
Joined
Apr 3, 2020
Messages
323
Reaction score
412
Location
Bay Area, California
Rating - 0%
0   0   0
Yeah, I saw several threads on this. The probes do get precipitant, and sometimes they even get those little tube worm looking things in them. Although on the threads some claimed actual probe failure or pm2/brain failure. So if cleaning it did not fix the problem, it looks like it's either the probe or module/brain the probe is connected to.

I will try to find some time to do a little research on this, and type up Troubleshooting section. If you can remember any titles/keywords from the threads you read, sharing them here (or even better, links to them) would be super helpful. Thank you for this!!
 

chipmunkofdoom2

Always Making Something
View Badges
Joined
Jun 6, 2017
Messages
2,417
Reaction score
4,505
Location
Baltimore, MD
Rating - 0%
0   0   0
No wonder most people think the salinity probe is junk. There's a "gotcha" where you don't get TC if the TC factor is 0 when you install the salinity probe? TC is automatically set to zero because a non-zero value is unsafe if there's no temperature probe installed? The temperature probe and salinity probe absolutely must be installed on the same module, no exceptions? Am I understanding this correctly? That's kind of disgraceful.

I would expect this if we borrowed the Apex from the microbrewing or hydroponics industry and are trying to make it work in a use-case for which it wasn't designed. This is a controller that was (ostensibly) designed from the ground up to do nothing but monitor and control aquariums. It also retails for what is a mortgage/rent payment to a lot of people. You should be able to plug in the salinity probe, maybe tweak the TC and calibrate, then be done with it.
 
OP
OP
eag

eag

Active Member
View Badges
Joined
Apr 3, 2020
Messages
323
Reaction score
412
Location
Bay Area, California
Rating - 0%
0   0   0
No wonder most people think the salinity probe is junk. There's a "gotcha" where you don't get TC if the TC factor is 0 when you install the salinity probe? TC is automatically set to zero because a non-zero value is unsafe if there's no temperature probe installed? The temperature probe and salinity probe absolutely must be installed on the same module, no exceptions? Am I understanding this correctly?

Yes that is correct, and yes I also found it confusing, hence my search for answers. Frankly, I think Apex could make just two small changes that would vastly improve the experience: update the software so that setting TC in the absence of a temperature probe is a recoverable/handled condition (thus allowing it to be enabled by default), and add temperature-related text to the calibration wizard and documentation (to avoid this very common calibration mistake).
 